I already calculated protein concentration using Bradford estimation but I need to calculate molar extinction coefficient by beer Lambert law. will any one help me?
Estimating of the molar exctinction coefficient (I assume at 280 nm) and the concentration requires the knowledge of your protein (the best would be the amino acid composition). Then you can use some online tool to estimate the parameters (link bellow).
Here is a paper on the subject - http://onlinelibrary.wiley.com/doi/10.1002/pro.5560041120/pdf
I have used the classical method for a single subunit protein. For that, you would prepare a solution of the purified protein. Run an absorption spectrum and record the absorbance at 280 nm. Using the same solution or a carefully diluted solution of it (keeping track of the dilution), you will need to determine the molar concentration. The Bradford estimation is not good enough. It would be more accurate to perform an amino acid analysis - http://www.sigmaaldrich.com/analytical-chromatography/analytical-reagents/amino-acid-analysis.html - which involves complete hydrolysis of the protein, separation and quantification of amino acids relative to a quantitative standard that is commercially available (http://www.sigmaaldrich.com/analytical-chromatography/analytical-reagents/amino-acid-analysis.html). From there, you would calculate how many micromoles or nanomoles you have of the rarest amino acid, that is stable to the acid hydrolysis. Using your knowledge of the protein molecular weight, you can do the calculation to arrive at the number of micromoles or nanomoles of protein you have per ml of solution. At this point, you would be able to apply the Beer-Lambert law because you will have both the absorbance at 280 nm and the molar protein concentration.
The expasy site that Jiri mentions is a good one. Practically, you need to be sure that your protein is >95% pure. If you have a contaminant that has lots of tryptophans (or if your protein has none) your estimation will be off using the 280nM read off. Bradford is better in such a case. If Bradford and UV are within10-20% error range you have nothing to worry.
The theoretical calculation of extinction coefficient based on amino acid composition assumes that the protein is completely denatured in 6M guanidine hydrochloride.
To determine the extinction coefficient experimentally, just make a solution of the protein with a known concentration (it could be based on the Bradford assay result or some other measurement of concentration) and measure the absorbance with a spectrophotometer, noting down the pathlength of the cuvette, and selecting the wavelength with the highest absorbance (usually about 280 nm, but it could be a different wavelength if there is a bound cofactor, prosthetic group, or ligand). The extinction coefficient is the absorbance divided by the concentration and the pathlength, according to Beer's Law (epsilon = absorbance/concentration/pathlength). The units of extinction coefficients are usually M-1cm-1, but for proteins it is often more convenient to use (mg/ml)-1cm-1. If the molecular mass of the protein is known, divide mg/ml (-g/L) by mass in g/mole to get mole/L.
A few things to keep in mind:
The Bradford assay does not always give the true value of protein concentration because it depends to some extent on the amino acid composition, and your protein may have a different composition from the standard used in the assay (usually BSA). However, it can still be used for an operational measurement of concentration and extinction coefficient.
UV absorbance can be exaggerated by light scattering caused by aggregated protein, and also by anything bound to the protein that absorbs. Making the absorbance measurement in 6M guanidine HCl should take care of the aggregation problem, but won't help with bound ligands or prosthetic groups.
Try http://www.protparam.net. Will let you enter your protein sequence and OD280 reading and give you your protein concentration. Much faster than expasy prot param too