As I diluted my sample from 1 molar to 100 ppm for HPLC analysis, Now if i want to correct my results from 100 ppm to 1 molar what will be the analytical method for that?
What should be the method? For example I have diluted the analytes initially from 1 molar to 100 pm and i got peak area let say X now i want the results corresponding to 1 molar, What will be the calculation?
We have to know the 'Molecular Weight' of your molecule of interest in order to perform the calculation. I just gave you the equations necessary to do this calculation.
Actually, i know how to dilute my sample but my query is related to how to get HPLC corrected results i.e results on the basis of sample before dilution
More information is needed before I can respond adequately. Is your HPLC method using an external standard with your sample? If so, your HPLC software will be able to calculate your sample (unknown) concentration based on peak responses (sample and standard), the final concentration of your standard and the dilution factor of your sample.
Your software test method and report method should have entries for standard concentration and sample dilution factor. Calculation of the sample concentration is straight-forward:
See ICH Q2 and Q3 on the requirements that need to be validated and develop your method to meet them. The FDA issued a white paper on the chromatography requirements in 1994. Take some courses from chromacademy.com !
How to dilute a sample if its initial concentration is not known, for example i am having an extraction sample of diethylamine in water but i don't know the concentration of diethylamine in the solution so if i want to dilute it for further analysis then what would be the procedure?
Typically, as part of method validation as per ICH Q2, you have already demonstrated a positive relationship (linear most commonly) between mass (as concentration, whether mg/L or ug/mL..) and peak response. Therefore;
(total volume of sample)/(total volume of aliquot) x concentration = mass