Sir please tell me what is thr meaning of this factor A(1%, 1cm)? Do we have to consider the dilution also ( if there is any) along with this calculation? And what is 'b' and 'c' in your abhove equation.? After measuring the absorbace how to use this factor along with the absorbance and dilutions to arrive at the result??
It is a value of typical sensitivity of a given analyte determined by a specific method. It does not mean that you don't need to calibrate your instrument, but often one needs to have some rough idea on sensitivity to check the method or to optimize sample preparation. In QC, high deviations from typical values may indicate some matrix or contamination problem, or some problem with the instrument. So I would recommend to calibrate your instrument according to your assay and to calculate the concentration needed for 1% absorption from your calibration curve. This should not differ too much from the typical value.
Dear Ashray, thank you for telling by e-mail that you have no standards for calibration, but that you need to rely in the extinction coeffcient. If you use 1 cm cuvettes c% is calculated as measured absorbance, divided by A(1%, 1 cm).